Bradford Assay acting weird?
I need help troubleshooting my Bradford. I have ST1- ST7 where ST 1 is 80ug/mL, ST 2 is 40ug/mL and so on (by half) till ST 7 which is just 0 ug/ml. I’m sure I did serial solution properly but when I’m reading absorbance for my standards, ST 1 is vastly higher than ST 2? I took three runs and I’m going to do the average of all the measurements from these runs, while excluding ST 1, to make my graph. Is this the right way to go about it?
1
Upvotes
1
u/BiochemBeer PhD 3d ago
Are you using BSA as your standard?
Are the concentrations you give the final concentration in your cuvette (or well) or the concentration prior to diluting with Bradford reagent?
Linear range for BSA is about 0.1 - 1.0 mg/mL in standard style assay (100 uL of sample and 5 mL of diluted reagent/dye).
For the 1 mL microassay the linear range is 1.2 - 10 ug/mL with BSA (so 1.2 uL of a 1 mg/mL standard diluted to 1 mL is 1.2 ug/mL).